Safety: Torque due gradients and main field

Post by motoca » Tue Nov 05, 2013 3:20 pm

Hi all,
I know that if you put a ferromagnetic material inside a MR bore, it suffers a torque or missile effect.
I know too that gradients can cause peripheral nerve stimulation and RF pulses are responsibile mainly to "heat tissues".Gradients are about miliTesla while B0 is about Tesla.
OK, after a brief introduction about I know about MR safety, I have "two" doubts:

1) Is there any torque due gradients field too? Or B0 is so huge that the gradients can't apply a additional shift in those materials? If they cause torque, Have I worry about it in in relation to patient safety as I have to have with B0 torques?
2) Do gradients cause any (a little) heat in tissues too or only RF pulses cause it?
Thanks in advance
Motoca

Dave Higgins
Site Admin
Posts: 522
Joined: Fri Sep 02, 2005 2:21 pm
Location: Leeds, UK
Contact:

Re: Safety: Torque due gradients and main field

Post by Dave Higgins » Thu Dec 19, 2013 3:53 pm

For practical purposes, you can ignore gradient fields for torque and missile effects.
Similarly for heating effects, the RF field need only be considered.

Dave