Sources of Image Noise

Answer this...

What are the two largest sources of noise (in images) in MRI?

Electronic noise generated in the receiver circuit and the patient (Brownian motion of ions in bodily fluids).

A modern scanner can keep the pre-amplifier noise down to well below 1 dB, and the use of local receive coils serves to increase SNR because (naturally) the signals received in the region of the surface coil have a higher amplitude. In quadrature coil design, the two receivers are at 90° to one another, enabling the distinguishing of the real and imaginary components of the received signal, increasing the SNR by √2. RF shielding keeps out anomalous external signals, and so they are not considered here.

Further reading on this topic:
Books: MRI From Picture to Proton p66, Q&A in MRI p63-65, MRI The Basics p28
Online: Basics of MRI